Practical Cryogenics
An Introduction to Laboratory Cryogenics
By N H Balshaw
Published by Oxford Instruments Superconductivity Limited
Old Station Way, Eynsham,
Witney, Oxon, OX29 4TL, England,
Telephone: (01865) 882855
Fax: (01865) 881567
Telex: 83413

Page 2

Page 3
Contents
1
Foreword................................................................................................................... 5
2
Vacuum equipment .................................................................................................. 7
2.1
Vacuum pumps ........................................................................................... 7
2.2
Vacuum accessories .................................................................................. 12
3
Detecting vacuum leaks ......................................................................................... 14
3.1
Introduction.............................................................................................. 14
3.2
Leak testing a simple vessel ..................................................................... 15
3.3
Locating 'massive' leaks ........................................................................... 16
3.4
Leak testing sub-assemblies..................................................................... 17
3.5
Testing more complex systems ................................................................ 17
3.6
Leaks at 4.2 K and below ......................................................................... 19
3.7
Superfluid leaks (or superleaks)............................................................... 19
3.8
Overpressure leak detection.................................................................... 20
4
Cryostats and coolers.............................................................................................. 21
4.1
Bath cryostats............................................................................................ 21
4.2
Lambda point refrigerators ..................................................................... 24
4.3
Continuous flow cryostats........................................................................ 26
4.4
Static and dynamic continuous flow systems.......................................... 28
4.5
Storage/transport dewars ........................................................................ 29
4.6
Closed cycle coolers .................................................................................. 30
4.7
'Stinger' systems ....................................................................................... 30
4.8
Peltier effect coolers................................................................................. 30
4.9
Making indium seals................................................................................. 30
5
Ultra low temperatures .......................................................................................... 32
5.1
3
He Refrigerators ...................................................................................... 32
5.2
3
He/
4
He Dilution refrigerators.................................................................. 35
5.3
Sorption pumped dilution refrigerators ................................................. 38
5.4
Nuclear demagnetisation systems ........................................................... 38
6
Superconducting magnet technology ................................................................... 39
6.1
Introduction.............................................................................................. 39
6.2
Construction of the magnet .................................................................... 40
6.3
Basic physics of the magnet..................................................................... 41
6.4
Homogeneity of the field ........................................................................ 42
6.5
Persistent mode operation....................................................................... 42
6.6
Quenches................................................................................................... 43
6.7
Protection circuit ...................................................................................... 44
6.8
Magnet power supplies............................................................................ 45
6.9
Typical operating procedure.................................................................... 46
6.10
Magnets for special applications............................................................. 47
6.11
Interfacing superconducting magnets to dilution refrigerator
systems ...................................................................................................... 48
6.12
Summary ................................................................................................... 49

Page 4
7
Transferring cryogens............................................................................................. 50
7.1
Liquid nitrogen......................................................................................... 50
7.2
Liquid helium............................................................................................ 50
7.3
Using liquid helium efficiently................................................................. 51
7.4
Avoiding helium transfer problems ........................................................ 53
7.5
Common problems ................................................................................... 53
8
Cryostat wiring........................................................................................................ 56
8.1
Thermal requirements.............................................................................. 56
8.2
Electrical requirements............................................................................. 57
8.3
Practical techniques.................................................................................. 58
8.4
Ultra low temperatures............................................................................ 61
8.5
UHV systems.............................................................................................. 61
9
Properties of materials ........................................................................................... 62
9.1
Physical properties of helium and nitrogen............................................ 62
9.2
Thermal conductivity integrals ................................................................ 62
10
Useful formulae and information.......................................................................... 67
10.1
Thermal conductivity and gas cooling .................................................... 67
10.2
Thermal radiation..................................................................................... 68
10.3
Convection................................................................................................ 68
10.4
Cooling materials to 4.2 K using liquid helium ...................................... 69
10.5
Superconducting transitions of common materials................................ 69
11
Glossary of terms .................................................................................................... 70
12
Useful reference books........................................................................................... 95
12.1
General practical techniques ................................................................... 95
12.2
Safety ........................................................................................................ 96
12.3
Thermometry and instrumentation......................................................... 96
12.4
Properties of materials............................................................................. 96
12.5
Theoretical reference books .................................................................... 96
 Oxford Instruments Superconductivity Limited, 1996-2001. All rights strictly reserved.
ISBN 0 9527594 0 3
This document is intended to be used only as background information, and not
as an instruction manual. No liability will be accepted for any damages
incurred because of the information contained in, or implied by, the contents.
First edition published 1996.
This imprint 2001.

Page 5
5
1
Foreword
This booklet has been written to help you to learn more about the basic principles of
cryogenics, so that you can design your experiments to make the best possible use of your
system. A little training often makes the difference between success and failure for a low
temperature experiment.
The booklet is a collection of practical notes to introduce beginners to the fundamentals
of good practice. It contains a 'glossary of terms' to explain some of the jargon commonly
used in cryogenics. The descriptions are intended to translate these terms into plain
English so that beginners can understand them.
The other sections give general advice on a range of relevant topics. A strong emphasis is
placed on practical information rather than theoretical details. Previous editions of this
booklet (then called Elementary Practical Cryogenics) contained some of this information.
Several small errors have been corrected and more information has been added.
The subject of 'safety' has deliberately been omitted. All cryogens are potentially
hazardous. Before you try to use a cryogenic or high magnetic field system you should
receive training from a competent person who knows your laboratory and the laws in your
country. You may then like to use the booklet Safety Matters (available from Oxford
Instruments) to remind you about this training when you are using a system.
N H Balshaw

Page 6

Page 7
7
2
Vacuum equipment
Vacuum systems are used most commonly in laboratory scale cryostats and
superconducting magnet systems for the following purposes:
To pump out the high vacuum insulation spaces in the cryostat and transfer tube
To pump out an exchange gas
To set up a pressure gradient along a pumping line so that the flow of cryogen
through the cryostat can be controlled
To reduce the vapour pressure over liquid helium surfaces where temperatures below
4.2 K are required
To pump out the nitrogen gas from a pre-cooled helium vessel, after the liquid has
been blown out
All gases, except helium, hydrogen and neon, will condense on surfaces cooled to below
about 60 K. Therefore, once liquid helium at 4.2 K is introduced into a vacuum vessel, all
the residual gases that are normally present will condense (or cryopump), reducing the
pressure in the vacuum space by one or two orders of magnitude. Therefore, the function
of a vacuum system is to reduce the pressure in the vacuum space to a point where the
thermal insulation is sufficiently good to allow liquid helium to be held in the vessel. In a
typical laboratory scale system the pressure then drops to 10
-5
mbar or less.
In cryostats that contain only liquid nitrogen, the coldest surface is at 77 K, which is above
the temperature for effective cryopumping by a metal surface. If the cryostat is not
pumped continuously by an external pumping system, a sorption pump is mounted on the
liquid nitrogen reservoir to maintain the integrity of the vacuum. Occasionally it has to be
cleaned by warming it to a temperature around 100
o
C and pumping the vacuum space. It
pumps air to a very low pressure when cooled with liquid nitrogen.
A booklet is available from Leybold to describe how to do most common vacuum
calculations. (See section 12).
2.1
Vacuum pumps
2.1.1
Single stage rotary pumps
Rotary pumps are used as roughing pumps (to reduce the pressure to a rough vacuum) or
as backing pumps (with a diffusion pump or turbomolecular pump). If the rotary pump's
sole function is to back a small oil diffusion pump or a turbomolecular pump then a single
stage rotary pump with a base pressure of about 10
-2
mbar and a displacement of about
5 m
3
/hour is adequate. However, some laboratories prefer to use a two stage rotary pump
with a diffusion pump because of the risk of stalling the diffusion pump if the backing
pressure exceeds a critical value (about 10
-1
mbar). If the diffusion pump stalls oil back-
streams into the vacuum system and can permanently affect the performance of the
cryostat.
If the rotary pump is also to be used as a roughing pump or to reduce the vapour pressure
over a liquid surface it may be necessary to choose a higher displacement pump to suit the
requirement. Most vacuum equipment manufacturers supply the information needed to
calculate the pump size in their sales brochures.

Page 8
8
In most cases, it is best to use a pump fitted with a 'gas ballast' facility. This helps the
pump to remove condensable vapours from the vacuum space of a cryostat. It is common
for water to accumulate in the vacuum spaces of cryostats if the cold surfaces are ever
exposed to air or if the cryostat is left unused for some time. Most surfaces release
absorbed water vapour when the pressure is reduced.
Practical base pressure:
10
-2
to 10
-3
mbar
Max. working pressure:
1 bar (for a limited period), few hundred mbar
continuously
Ideal for:
Rough pumping, backing high vacuum pumps, lambda
point refrigerators, variable temperature inserts, 1 K pots
Figure 1 Rotary pump system
2.1.2
Two stage rotary pumps
In some cases, it is possible to replace a rotary / diffusion pump combination with a two
stage rotary pump. A base pressure of 10
-4
mbar can be achieved in ideal conditions and
using a cold trap, but in practice the pressure in the cryostat will probably only be 10
-1
to
10
-2
mbar.
This type of pumping system is very simple but it cannot reach a low enough base pressure
to give good thermal isolation . A large amount of liquid helium would be required to
cryopump the residual gas and the static boil off of the system would be slightly increased.
In addition, if the cryostat is used above 60 K and there is no sorption pump in the vacuum
space, condensation or frosting may be seen on the outside of the cryostat.
2.1.3
Diffusion pumps
A 50 to 75 mm diameter oil diffusion pump is sufficient for pumping laboratory scale
cryostats. An air-cooled pump with an air pumping speed of about 50 litres/second and an
ideal ultimate pressure of about 10
-7
mbar is commonly used.

Page 9
9
It is advisable to use a cold trap with a diffusion pump (although some people do not
consider it to be essential). You should never pump the vacuum space of a cold cryostat
without a cold trap between the pump and cryostat. This trap helps to remove water
vapour from the vacuum space and prevents back streaming of oil vapour from the pump.
Two types of trap are commonly used, liquid nitrogen filled traps and thermo-electric
(Peltier effect) cooled baffles. The latter require less attention and are better for very
long-term unattended operation.
Practical base pressure:
10
-7
mbar
Max. working pressure:
10
-2
to 10
-1
mbar
Ideal for:
Pumping insulating high vacuum spaces in cryostats (e.g.
OVCs)
Figure 2 Diffusion pump system
2.1.4
Turbomolecular pumps
These high vacuum mechanical pumps can be used instead of diffusion pumps. They are
especially useful if a very clean high vacuum is needed because the compression ratio is
strongly dependent on the mass of the molecules. The large hydrocarbon molecules are
pumped so well that there is virtually no backstreaming of oil. Many of the modern
pumps incorporate a molecular drag stage within the pump and can tolerate a backing
pressure of 10 mbar or higher. An oil free diaphragm pump can then be used as the
backing pump for some applications.

Page 10
10
Practical base pressure:
10
-8
mbar
Max. working pressure:
A few mbar (for conventional turbomolecular pumps).
About 30 mbar (for some pumps with a molecular drag
stage).
Ideal for:
Pumping clean high vacuum spaces (with or without cold
trap).
Figure 3 Turbomolecular pump system
The pumping speed for helium is about 20% higher than that for nitrogen, but the
compression ratio is much lower. Therefore if the pump is to be used to pump helium
from a vacuum space it is best to use a two stage rotary pump, so that the backing
pressure is as low as possible. A diffusion pump is still better at pumping helium!
Turbomolecular pumps should only be vented (from the high vacuum side) while they are
still spinning slowly. This helps to prevent contamination backstreaming from the high
pressure side of the pump. Most pumps can be fitted with an automatic venting device
which is activated by the pump controller. The gas can be drawn through a drier cartridge
to prevent contamination with water.

Page 11
11
2.1.5
Roots pumps
Roots pumps (or roots blowers) are mechanical booster pumps, used (in conjunction with a
backing pump) to reach the medium to high vacuum range with very high gas
throughputs. Two (or more) pumps can be used in series with some advantage. For
example, if you need a pumping speed of 1000 m
3
/h it may be best to use a 1000 m
3
/h
roots pump backed by a 250 m
3
/h roots pump, which in turn is backed by a 65 m
3
/h rotary
pump. Vacuum companies often recommend a 1000 m
3
/h roots pump backed by a
250 m
3
/h rotary pump, but this option is usually more expensive.
Practical base pressure:
10
-4
mbar
Max. working pressure:
Few hundred mbar
Ideal for:
High volume flow rates in the pressure range 10
-4
to 50 mbar
(for example, in dilution refrigerator systems)
Figure 4 Roots pump system
2.1.6
Sorption pumps (or sorbs)
Sorption pumps are often used in vacuum spaces because they are cheap and reliable, and
they require little maintenance. The adsorbent material (usually activated charcoal or a
molecular sieve) has a very large surface area, and the gas molecules are trapped onto the
surfaces when the sorb is cold.
Liquid nitrogen cryostats usually have a sorb fitted to the outside of the nitrogen vessel to
maintain a good insulating vacuum. If a sorb is not used the vacuum slowly deteriorates
as the warm surfaces outgas. A 77 K sorb will not trap helium gas, but if it is cooled to
4.2 K helium gas may be pumped to below 10
-5
mbar. Liquid helium vessels are cold
enough to freeze any gas except helium onto the metal surfaces, so a sorb is not usually
fitted because it may hinder leak detecting operations.

Page 12
12
These pumps are single shot devices. They eventually become saturated and have to be
warmed (and sometimes evacuated with a suitable high vacuum pumping system) to
regenerate the absorbent material. The amount of gas that can be pumped before the
sorb is saturated depends on the type of gas and the temperature of the pump, but in a
high vacuum environment, they may be expected to last for a period of months or years
before they need to be regenerated.
2.1.7
Cryopumps
A cryopump usually consists of a large number of metal plates cooled to a temperature
close to 4.2 K, (either by liquid helium or by a closed cycle cooler). Like sorption pumps,
these are single shot pumps and they have to be regenerated when a layer of ice has
collected on the metal surfaces. The pump relies on the fact that the vapour pressure of
most materials at a temperature below 10 K is negligible. This type of pump is essentially
clean and it is suitable for use in ultra high vacuum systems.
2.2
Vacuum accessories
2.2.1
Oil mist filter
An oil mist filter is used to remove the fine mist of oil from the exhaust gas of a rotary
pump. It is desirable to remove this mist for the following reasons.
The vapour represents a health hazard if inhaled
It may contaminate any flow meters or other fittings behind the pump, changing their
calibration
It is desirable to avoid contaminating the helium recovery system with pump oil.
Several different types of filter are available. The following are the most common:
a) Coalescing filters, which only need to be replaced if they are dirty. The oil normally
runs back into the pump continuously.
b) Centrifugal filters or 'catch pots', which usually have a transparent bowl to collect the
oil, and have to be emptied occasionally.
2.2.2
Vacuum gauges
High vacuum is normally measured using a combination of Pirani and Penning type
gauges. Typically, the Pirani gauge operates in the range 10 mbar to 10
-3
mbar, and the
Penning in the range 10
-2
mbar to 10
-7
mbar. The calibration of these gauges (and some
others) depends on the type of gas in the system.
Vacuum in the range from 1 to 1000 mbar is normally measured with reasonable accuracy
using a simple capsule or dial gauge. Other types of gauge are available, for example,
Piezoelectric gauges and Baratron gauges, which allow accurate remote measurement.
2.2.3
Pumping lines
The pumping lines may have as large an effect on the efficiency of the vacuum system as
the pumps themselves. Check the following points:
a) The lines must be leak tight. Plastic or rubber materials are sometimes used as
pumping lines but they may be permeable to helium gas.

Page 13
13
b) The throughput of the lines must be at least as high as that of the pumps. Otherwise
their impedance limits the flow of gas and may affect the base pressure at the cryostat
end of the line. It will certainly affect the amount of time required to pump down to
the required pressure.
c) The lines should be clean inside. If there is any moisture in the lines it will limit the
pressure that can be reached. If the lines are heavily contaminated with helium gas it
will be difficult to perform the normal leak tests.
2.2.4
Mass spectrometer leak detectors
Although these machines are expensive, it is very useful to have access to one. The leak
detector need not be dedicated to one system; it can be shared by the lab or the
department. In general, the more complicated the system is, the more useful the leak
detector will be. For example, a complex dilution refrigerator system may have 500 to
2000 joints that must be leak tight. Many of these are subject to thermal cycling, and a
leak from almost any of them could cause a system failure. It is clearly important that any
leaks can be traced and cured as quickly and easily as possible. A sensitivity of 10
-8
mbar l/s
(or standard cm
3
/s) is sufficient for most purposes, but 10
-10
mbar l/s is preferred when
looking for very fine leaks or superleaks.
If you have little or no experience of using these leak detectors , refer to section 3.
2.2.5
Foreline traps
A foreline trap is sometimes used on the inlet of a rotary pump to reduce the amount of
oil backstreaming up the pumping line. The active material in the foreline trap must be
changed regularly so that it remains effective.
2.2.6
Choosing an appropriate 'O' ring material
Silicone rubber is often used for 'O' rings in electrical equipment, but it is not generally
suitable for cryogenic equipment because the material is porous to helium gas. However,
it is probably suitable for dynamic seals at temperatures down to -60°C, or static seals
down to -100°C. It can also be used up to 250°C.
Butyl rubber was an old favourite material for vacuum applications, and it was often used
because of its low gas permeability. Suitable for temperatures down to -60°C.
Nitrile rubber (or Buna N) is probably the best material for most common vacuum
applications. It is cheap, easily available in a range of sizes, and appropriate for
temperatures slightly below room temperature. It is also resistant to silicon grease (for
example, vacuum grease). Its working temperature range is from -40°C to +120°C.
Fluoroelastomer (for example, 'Viton') is also suitable for vacuum. It is better than nitrile
rubber for high temperature applications, but it is more expensive and tends to be
deformed permanently after being compressed for a length of time. Its working
temperature range is from -20°C to +200°C. Beware: if it is subjected to temperatures
much higher than 200
o
C, the black sticky residue contains hydrofluoric acid!
Teflon (or PTFE) can also be used to make vacuum seals. However, the joint has to be
designed to prevent the Teflon 'creeping' when it is compressed.

Page 14
14
3
Detecting vacuum leaks
3.1
Introduction
These notes describe how to locate leaks in complex vacuum systems using a helium
sensitive mass spectrometer leak detector. They do not describe how to use the leak
detector in detail, because so many different models are available. Consult the instruction
booklet for this information (and good luck!). Better still, ask someone to show you how
to use the leak detector.
Warning:
Before you attempt to carry out a leak test, it is important to check that it is safe
to evacuate a vessel, and that there is no risk of it collapsing because of the
external pressure of the atmosphere. This is especially important for vessels
which have thin walled tubes (for thermal reasons) and for large vessels. If you
collapse a vacuum vessel you might be badly injured by the shock wave or by
flying fragments.
Helium sensitive leak detectors are used because:
Helium atoms are small and mobile, so they can pass through small holes easily
Helium gas is inert and safe to use
There is very little helium in the air allowing the leak to be located precisely
Vacuum leaks are most commonly associated with:
Welds
Leaks caused by imperfect welds, cracked welds, or corrosion around the weld.
Soldered joints
Leaks caused by imperfect joints, or corrosion.
'O' ring seals
Dry, damaged or broken 'O' rings, or scratches or hairs lying across the seal are the most
common sources of problems. (As a rough guide, a hair lying across an 'O' ring may cause
a leak in the range 10
-6
to 10
-3
mbar ls
-1
, depending on many factors.)
Indium seals
Insufficient compression of the indium wire, dirt on the metal faces or scratches across the
seal may cause leaks. Problems after thermal cycling might point to poor flange design.
Glued joints
Leaks may be caused by bad joint design, bad surface preparation, inappropriate choice of
glue, or rapid thermal cycling.
Porosity of metals
Gas sometimes leaks along the grain of a metal (especially in some grades of brass). Small
flanges are commonly made of plate rather than bar for this reason.
Diffusion
Many plastics and composites are porous to helium gas at room temperature but not
when cooled down. Materials must be chosen appropriately for the working environment
and temperature range.
Thermal cycling
Leaks may be undetectable at room temperature but only open when the component is
cooled to liquid nitrogen temperature. Sometimes the leak will still be detectable when it
is warmed up again. If not, repeated thermal cycling may help to open the leak path and
make detection easier.

Page 15
15
Superfluid leaks
Components that are leak tight at room temperature and even in liquid helium may leak
when subjected to superfluid helium (which has zero viscosity). This is the most difficult
type of leak to find!
3.1.1
Getting started
Vacuum leak detection is an art, but a scientific approach helps. When you start to learn
how to use the leak detector you will almost certainly find yourself looking for leaks that
do not exist, and you could waste hours if you are not careful. These notes should help
you to avoid most of the common problems.
Most leak detectors have an audible signal and a visual display. Both of these are useful.
The visual display is used to quantify a leak and detect a slow change in the signal, so it is
especially useful to help you locate small leaks. The audible signal is much easier to use
for general leak testing, because you do not have to look at the leak detector. You can
then concentrate on looking at the equipment that you are testing and if you hear the
signal rise you can go back over the same area again more slowly, and try to pin point the
position of the leak.
From time to time the sensitivity of the leak detector should be checked and reset using a
'standard leak', since the sensitivity peak may drift.
3.2
Leak testing a simple vessel
3.2.1
Preparations
Consider first how to test a simple vessel for leaks: for example, a flexible pumping line.
The principles learnt here can then be extended to more complex systems.
Evacuate the line to a rough vacuum using a suitable rotary pump, and then pump it to a
sufficiently high vacuum for the mass spectrometer to be used (typically 10
-5
mbar). Many
leak detectors will evacuate the vessel and switch on the mass spectrometer for you
automatically.
Select a suitable sensitivity range so that a small leak can be detected. For most cryogenic
systems the 10
-8
mbar l/s range (or 10
-8
standard cm
3
/s) is best. If you use a more sensitive
range than this the background helium signal in the vacuum space may exceed full scale
on the leak detector. If you use a less sensitive range you may not notice the leak.
If the background signal is too high to allow you to use a sensitive range, pump the vessel
until the signal has been reduced sufficiently. You can sometimes reduce the signal more
quickly by 'pumping and flushing'. Pump the air out of the system until it reaches a
pressure of a few mbar, allow dry nitrogen (or air if this is not available) into the vacuum
space again (slowly to avoid damaging the vacuum vessel), and repeat the process as often
as necessary.
3.2.2
Leak testing the pumping line
Spray the pumping line with helium gas, paying special attention to any joints. If possible,
place it in a plastic bag, and fill the bag with helium, so that there is no chance of missing
a small leak in an unexpected position. However this is likely to be impractical for large
vessels.

Page 16
16
If the signal on the leak detector rises at any time during the test, a leak should be
suspected, and you should methodically check to find out whether the leak is real, or an
artefact caused by the outgassing of some trapped gas within the vacuum system.
3.2.3
Work from the top
Remember that helium gas is lighter than air so it rises. Therefore you should start by
spraying gas on the highest point, and slowly work downwards. If the signal on the leak
detector rises at any time, go back over the area that you have just covered, and check
again. If you do not start at the top you can get misleading results when you are checking
an area below the position of the real leak.
When you have found the approximate location of the leak, fit a fine nozzle to the end of
the helium gas line, and reduce the flow of gas. Check the suspect areas in detail. You
can locate leaks very precisely. Usually (but not always) you can see a small hole, flaw in
the material or dullness of the surface at the leak position.
If you want to test a long weld on a large vessel you can fix a tunnel of plastic sheet to the
vessel with adhesive tape and fill the tunnel with helium gas.
3.3
Locating 'massive' leaks
Occasionally you may find a leak that is so big that you cannot reduce the pressure
sufficiently to use the mass spectrometer. How can you find the position of the leak?
3.3.1
The safe method
Connect a large displacement medium vacuum pumping system in parallel with the mass
spectrometer, as shown in Figure 5.
Figure 5 Locating a massive leak in a vacuum system
Pump the vacuum system to a rough vacuum with the rotary pump. Then slightly open
the throttle valve on the pumping port of the mass spectrometer. Make sure that the
pressure at the inlet of the leak detector is not too high for it to work properly.

Page 17
17
It is sometimes possible to check for a leak by slightly pressurising the vessel with helium
gas as described in section 3.8.
3.3.2
The other way (at your own risk)
Some people use water, acetone or methanol to locate leaks on small systems. Open the
gas ballast valve on the pump to make sure that contamination does not collect in the
pump oil.
Hazard:
Acetone or methanol are flammable so you must not use them in large quantities. Take
care not to create a fire hazard - contact your safety officer first.
Pump the vessel to a rough vacuum and measure the pressure. Brush liquid onto the
outside of the vessel. The pressure rises quickly when the area of the leak is found.
These liquids may also be used to block a 'massive' leak temporarily, so that the rest of the
vessel may be tested. Apply the liquid with a brush. Initially the pressure rises, but soon
the liquid freezes and blocks the leak. You can remove the ice by gently warming the
area with a hot air blower.
3.4
Leak testing sub-assemblies
If you are building a complex system you can test the sub-assemblies before you join them
together. In this way you can locate leaks before the system is assembled. Components
that are not fitted with standard vacuum fittings can be sealed to suitable plates using a
product such as Apiezon
TM
'Q compound'. This is a malleable material that can be used to
make a temporary seal, but it is only suitable for use at room temperature.
Occasionally you may find a component that has a detectable leak in one direction but not
in the other. Therefore it is best to test components by evacuating the side that will be
under vacuum in the finished assembly. In any case you must check that it is safe to
evacuate the vessel, and that there is no danger of it collapsing.
3.5
Testing more complex systems
Most real cryogenic systems are quite complex. It may be useful to consider a liquid
nitrogen shielded liquid helium dewar, which would usually be tested at room
temperature and at liquid nitrogen temperature. If a vessel is leak tight at 77 K it is
unusual for it to develop a leak as it is filled with liquid helium at its normal boiling point.
This is probably because most materials undergo very little thermal contraction below
77 K, so thermal stresses induced by cooling to 4.2 K are smaller than those caused by the
initial pre-cooling process.
The dewar must be leak tight in the following ways:
Outer vacuum chamber (OVC) to air
Liquid helium reservoir to OVC
Liquid nitrogen reservoir to OVC

Page 18
18
Evacuate the OVC, and set up the leak detector to monitor it. Check the outside of the
dewar as described in section 3.2. Then flush the liquid nitrogen and liquid helium vessels
with helium gas. The best way to check them thoroughly is to pump the air out of each
vessel using a small rotary pump, and then to fill each vessel in turn with helium gas.
Sometimes the signal on the leak detector rises and falls again as the pressure in one of
the reservoirs changes. This might not indicate a leak; small movements of the vessels can
release gas from the surfaces. You can check this by repeating the test using air instead of
helium gas.
If there is any doubt about the presence of a minute leak, use a chart recorder to monitor
the signal from the leak detector. You should then see a step on the chart when the
helium gas is allowed into (or pumped out of) the suspect space. This makes it easier to
distinguish between a real leak signal and noise on the signal.
When you have finished the room temperature leak tests, you can pre-cool the cryostat to
77 K. Then blow the liquid nitrogen out of each space in turn using helium gas, while you
monitor the OVC with the leak detector.
3.5.1
Pumping and flushing with helium gas
When you have removed all of the liquid, pump the vessel to a pressure of a few mbar,
and then fill it with helium gas. This ensures that small leaks are not blocked by
remaining droplets of liquid nitrogen.
3.5.2
Temperature effects
If you see a signal rise it may have been caused by a temperature change. Helium gas
(trapped on the surfaces) may be released as warm gas is allowed into the vessel.
3.5.3
Masking cold leaks
If you discover a cold leak, you might be able to determine the approximate position of
the leak by refilling the vessel with liquid nitrogen and then blowing the liquid out again
with helium gas. When you see the signal rise again stop blowing out the liquid, and
measure the level of the liquid. The leak is probably at this height in the vessel.
3.5.4
More complicated systems
You can check even more complicated systems in a similar way; for example, dewars with
variable temperature inserts or dilution refrigerator inserts in the liquid helium reservoir.
Think in advance about the best order to carry out the leak tests. In this way you can
often reduce the number of pumping operations. You may be able to test several vessels
at once. It is only necessary to check them individually if you discover a leak.

Page 19
19
3.6
Leaks at 4.2 K and below
Any leak found at a temperature below 4.2 K is difficult to locate, because helium gas
tends to be absorbed onto the cold surfaces. It may take a long time for the leak detector
to respond to a small leak, so patience is essential. Do not be tempted to hurry the tests.
It may take many hours for a positive result to be obtained. It is sometimes best to warm
up the system to 77 K and pump the helium gas away thoroughly before cooling to 4.2 K
again, because this removes the gas absorbed on the cold surfaces. Check each possible
source of the leak in turn, with the others under vacuum to eliminate any possibility of
confusion. At best, you will only be able to determine which space is responsible for the
leak.
If the leak is very small, it is possible for the leak detector to pump away the helium gas at
the same rate as the leak, and so the signal may not be seen to rise at all. In this case,
close off the vacuum space for a few hours, and compare the signals before and after the
test period. Make sure that the conditions for both readings are identical, so that thermal
effects on the outgassing rate can be neglected.
3.7
Superfluid leaks (or superleaks)
Superleaks will only be seen at temperatures below the lambda point (2.2 K). Superfluid
liquid
4
He (also known as helium II) has zero viscosity and can pass through very small
holes quickly. Fortunately these leaks are quite rare, because it can take days or weeks to
cure them.
Location of the precise position of the leak is extremely difficult. It may be possible to
open up the leak sufficiently to detect it at 77 K (or even room temperature) by rapid
thermal cycling, but this technique is not always successful. Failing this, the only other
course of action is to replace joints or components, starting from the easiest operations.
This would normally be done in the following order (probably with a leak test after each
step):
Indium seals replaced
Wood's metal joints re-run
Soft soldered joints re-run
Silver soldered joints re-run
1
Welds re-made
Components or sub-assemblies replaced
1
You can only re-run silver soldered joints if there is no soft solder nearby. If soft solder is heated to
the melting temperature of silver solder it can dissolve other materials into solution. It is then almost
inevitable that the joint will leak, even if a large hole has not appeared.

Page 20
20
3.8
Overpressure leak detection
Overpressure helium sensitive leak detectors ('sniffers') can be used to detect helium gas in
the air. You can use these to detect very large leaks on vacuum systems but they should
not be relied upon for the routine testing of cryogenic equipment. The vessel should be
slightly pressurised with helium gas, and the sniffer is then used to detect where the gas is
escaping. It is also important to check that it is safe to over-pressurise the vessel before
trying to use this technique.
Sniffers are especially useful to detect the location of a leak in a helium recovery system,
or on the (room temperature) fittings on the helium reservoir of a cryostat.

Page 21
21
4
Cryostats and coolers
Various types of cryostat are available, and each type has advantages and disadvantages
compared with the others. The following notes may help you to decide on the best type
of system for your application.
In general, if the cryostat has to be very large (for example, to contain a superconducting
magnet or conventional
3
He/
4
He dilution refrigerator) it is best to use a 'bath' cryostat.
You can fit a suitable continuous flow insert within the bath cryostat to achieve the
sample temperature range required for your experiment.
However, if the cryostat has to fit into a small space or has to be thermally cycled rapidly
and often, and the experimental equipment does not need a self contained reservoir of
cryogen, it may be better to feed liquid from a remote storage dewar through a special
transfer tube. This is called a 'continuous flow' cryostat.
The different types of cryostat that are widely available are described individually in the
following sub-sections.
4.1
Bath cryostats
Bath cryostats contain large enough supplies of cryogens for a convenient period of
operation. There is no need to refill the cryostat continuously from a storage dewar. The
'hold time' depends on a number of factors, (for example, size, experimental heat load
and cryogen consumption rate). They are typically designed to give operating periods
between 10 hours and 4 months.
Two types of bath cryostat are commonly used for laboratory scale liquid helium
temperature systems. Both types are vacuum insulated to reduce the heat load due to
conduction and convection. However, the helium reservoir is shielded from the room
temperature radiation heat load in different ways. According to Stefan's Law, the amount
of heat radiated from a warm body to a cold body varies with the difference between the
fourth power of their temperatures. Therefore a 300 K surface radiates 230 times more
heat to a 4.2 K surface than a 77 K surface would radiate onto the same 4.2 K surface.
Therefore liquid helium reservoirs are always shielded from room temperature radiation
by a cooled shield. In most cryostats, the radiation load is further reduced by the use of
'multi-layer superinsulation'. This consists of many thin layers of low emissivity material, in
the insulating vacuum space.
4.1.1
Liquid nitrogen shielded cryostats for liquid helium
In this type of cryostat, the liquid helium reservoir is surrounded either by a reservoir of
liquid nitrogen, or by a shield cooled by this reservoir. The liquid nitrogen vessel is
thermally linked to the neck of the liquid helium vessel to form a thermal barrier to heat
conducted down from room temperature. Figure 6 shows a typical small liquid nitrogen
shielded cryostat used to cool an infra-red detector to 1.5 K.

Page 22
22
Figure 6 Low loss infra-red detector cryostat
The advantages and disadvantages of liquid nitrogen shielded cryostats are summarised in
Table 1 on page 23.
4.1.2
Vapour shielded cryostats for liquid helium
As an alternative to liquid nitrogen cooled shields, it is possible to link several thermal
shields to the neck of the liquid helium vessel. The cold gas that has evaporated from the
reservoir is then used to cool these shields. This type of cryostat typically has between two
and six shields (depending on the required performance) linked to different points on the
neck. The space between the shields is filled with superinsulation.
The boil off rates of the two types of cryostat are similar, providing that there are no
dimensional constraints on the system. Table 2 shows the advantages and disadvantages
of vapour shielded cryostats.

Page 23
23
Advantages of liquid nitrogen
shielding
Disadvantages of liquid nitrogen
shielding
The shield forms a firm point to fix the
temperature of windows or thermal
anchors, and the temperature of the
shield is fairly constant.
The system may be warmed up quickly
by allowing gas into the vacuum space.
The small amount of superinsulation
does not become badly contaminated,
and the gas can be pumped out to an
acceptable level.
Comparatively short systems can be
made, because of the firm 77 K thermal
link in the neck of the helium vessel.
Liquid nitrogen must be filled regularly.
Boiling liquid nitrogen creates intermittent
vibration since it tends to boil in bursts. The
gas flow from the LN
2
exhaust port may be
very low for an extended period as liquid in
the cryostat stratifies. Liquid near the
bottom of the reservoir can become warmer
than the surface because of the hydrostatic
pressure of the liquid above it. When this
stratification is disturbed the evaporation
rate increases dramatically. This is sometimes
sufficient to blow liquid out of the cryostat.
Table 1 Liquid nitrogen shielded cryostats for liquid helium
Advantages of vapour shielded
cryostats.
Disadvantages of vapour shielded
cryostats.
Liquid nitrogen does not need to be re-
filled.
Vibration levels may be reduced, since
there is no vibration from the
intermittent boiling of liquid nitrogen.
Warming up the system may take longer
than it would take if the vacuum space could
be 'softened' with gas.
Very short systems may have a higher boil off
than a corresponding liquid nitrogen
shielded system.
The temperature of the shields varies with
the liquid helium level.
Table 2 Vapour shielded cryostats for liquid helium
4.1.3
Bath cryostats for liquid nitrogen
In some ways it is more difficult to make a reliable bath cryostat for liquid nitrogen than
for liquid helium. Unlike liquid helium, liquid nitrogen is not cold enough to freeze (or
cryopump) all the contaminating gases in the surrounding vacuum space onto a cold metal
surface. The quality of the vacuum is critical for operation of the cryostat, so a sorption
pump (containing charcoal or molecular sieve) is normally fitted to the outside of the
liquid nitrogen reservoir to maintain the vacuum. The vessel is usually superinsulated, but
other insulation techniques are occasionally used; for example, filling the vacuum space
with a low conductivity material such as a suitable mineral powder.

Page 24
24
4.2
Lambda point refrigerators
Superconducting magnets are usually operated in liquid helium at 4.2 K. Their
performance can often be enhanced by cooling the magnet to lower temperatures as
described in section 6.3 on page 41. The simplest way to achieve temperatures below
4.2 K is to pump the whole liquid helium reservoir with a rotary pump, to reduce the
vapour pressure above the liquid. If the bath is cooled to 2.2 K in this way, about 35% of
the helium is evaporated to cool the remaining liquid. Temperatures below 2.2 K can be
achieved, but if the bath is cooled below the lambda point, the liquid helium consumption
increases significantly (both to reach the low temperature and to maintain it).
This simple approach has several disadvantages. A large amount of liquid is used to cool
the magnet down, and since the reservoir is then below atmospheric pressure, access to
the reservoir is difficult and all the fittings on the top plate have to be reliably leak tight.
The liquid helium can only be re-filled by de-energising the magnet to its 4.2 K field and
filling the reservoir to atmospheric pressure with helium gas, which interrupts the
experiment.
Lambda point refrigerators (also known as 'lambda plates' or 'pumped plates') are used
to cool superconducting magnets to about 2.2 K and maintain this temperature
continuously. See Figure 7. They consist of a needle valve (to control the flow of liquid
helium into the refrigerator) and a tube or chamber with a pumping line. They are
normally built into the 'magnet support system'. The refrigerator is in good thermal
contact with the liquid helium just above the magnet.
Liquid is continuously fed into the refrigerator and pumped to a low pressure so that it
cools. The cooling power is determined by the liquid flow rate and the size of the pump,
and it can be adjusted using the needle valve. High flow rates are typically used at high
temperatures to cool the system quickly or to obtain high cooling power, but when base
temperature is reached, the flow can be reduced to make operation as economical as
possible.
The density of liquid helium changes rapidly with temperature, so strong convection
currents are set up, around the magnet. The cold liquid from the refrigerator sinks to the
bottom of the reservoir, cooling the magnet and keeping it at about 2.2 K. Meanwhile
the warmer liquid above the refrigerator is affected very little. The thermal conductivity
of the liquid is so low that the region immediately above the plate has a steep
temperature gradient, and the liquid surface remains at 4.2 K and at atmospheric pressure.
It is important to make sure that this thermal gradient is maintained, and not short
circuited by high conductivity components.

Page 25
25
Figure 7 Lambda point refrigerator
Lambda point refrigerators have several advantages. In particular:
a) Since only a small proportion of the liquid in the reservoir is cooled by the lambda
plate less liquid has to be used, and this reduces the cost of operation.
b) Operation can be automated (using a Teslatron Lambda controller).
c) The reservoir can be refilled without stopping operation of the system, as long as the
transfer tube does not stir the liquid and upset the temperature gradient above the
lambda plate.
The performance of these systems is dominated by the amount of liquid helium that has
to be cooled. Although the mass of the magnet is much larger than that of the liquid, its
heat capacity is very much lower. It is possible to calculate the amount of heat that has to
be removed if the magnet and liquid are cooled from 4.2 K to 2.2 K. In a typical system,
containing a 50 kg magnet, there may be about 3 litres (0.5 kg) of liquid below the
lambda plate. Only 5 J has to be removed from the magnet, but about 3 kJ has to be
removed from the liquid. Therefore it is important to minimise the amount of liquid
around the magnet so that it will cool quickly and cheaply.
In most systems the magnet can only be cooled to 2.2 K in this way, because liquid helium
has a phase change (the lambda point) at this temperature. Below the lambda point, the
liquid becomes 'superfluid' and has a very high thermal conductivity, so the phase
transition can only occur if the whole reservoir is cooled to the lambda point.

Page 26
26
The heat from any warmer region in the reservoir would be rapidly conducted to the
colder region, keeping its temperature above the critical level. However, in a few
specialised applications, the refrigerator is built into the top of a separate chamber around
the magnet. The refrigerator is fed from a 4.2 K liquid reservoir, but thermally isolated
from it. The lambda plate then cools the whole of this chamber, and temperatures below
the lambda point can be reached and maintained continuously, while the liquid is at
atmospheric pressure. The optimum temperature is about 1.8 K, as the superfluid is then
able to carry heat away from the magnet most effectively.
4.3
Continuous flow cryostats
A wide range of continuous flow cryostats is available. Some of these are supplied with
cryogens from a storage vessel; others are mounted in a bath cryostat which supplies
liquid. In most of these systems the cooling power available from a flow of cryogen (LN
2
or LHe) is balanced by power supplied electrically to a heater near the sample (usually by a
temperature controller).
4.3.1
Variable temperature inserts (VTI)
Variable temperature inserts are used in bath cryostats to adjust the temperature of a
sample without affecting the helium reservoir. 'Dynamic' and 'static' types of VTI are
available, and the advantages and disadvantages of each type are described in section 4.4.
The inner parts of the insert are vacuum insulated from the liquid helium. There may also
be a radiation shield between the sample space and the liquid reservoir to reduce the
radiated heat load on the reservoir when the sample is at a high temperature. This shield
is usually cooled by the exhaust gas or the boil off from the main bath.
The temperature range of a VTI is typically from 1.5 to 300 K, but in certain circumstances
this range may be extended. The sample temperature can be controlled continuously at
any point in this range. Lower temperatures can often be achieved in single shot mode:
the sample space is filled with liquid and the needle valve is closed to allow the pump to
reduce the vapour pressure above the liquid to the lowest possible level.
4.3.2
Independent continuous flow cryostats (CF)
The operating principles of CF cryostats are generally the same as those of VTIs. Dynamic
and static versions are available as described in section 4.4. However, they normally have
their own independent thermal shielding, and they are supplied with coolant from an
independent storage vessel through a 'low loss' or 'gas flow shielded' (GFS) transfer tube.
These cryostats are sometimes used with a superconducting magnet if it has a room
temperature bore. They are also used with resistive magnets.

Page 27
27
Figure 8 Optistat - an optical continuous flow cryostat
The transfer tubes are always vacuum insulated. In order to reduce the losses in the tube,
GFS type transfer tubes use the enthalpy in the exhaust gas from the CF cryostat to cool a
radiation shield in the tube.

Page 28
28
The temperature range of CF cryostats is typically < 4 to 300 K in continuous mode, with
lower temperatures available for limited periods in 'single shot mode'. However the
range may be extended to give higher or lower temperatures if necessary. In general it is
difficult to achieve temperatures as low as those available in VTIs because of the thermal
losses in the transfer tubes, but some cryostats are designed to reach 1.6 K continuously.
Figure 8 shows one of them schematically.
4.4
Static and dynamic continuous flow systems
Although all continuous flow cryostats work on the principle of balancing the cooling
power of a flow of cryogen with electrical power from the temperature controller, there
are several distinct types of cryostat: the most important are referred to as 'dynamic' and
'static'.
4.4.1
Dynamic systems
In a dynamic continuous flow cryostat, the sample is mounted in a flowing gas or in liquid,
and its temperature is strongly influenced by the fluid. The temperature of the fluid is
controlled by passing it through a heat exchanger (usually placed at the bottom of the
sample space). The heat exchanger temperature is set by simultaneously controlling the
cryogen flow rate and the heater on the heat exchanger. A temperature controller is
usually used to do this automatically. Providing that the flow of cryogen through the heat
exchanger is not too high the temperature of the flowing fluid can be controlled quite
accurately. The fluid flows past the sample and out of the exhaust port of the insert to
the pump.
This type of insert is easy to operate and it responds very quickly if the set temperature is
changed to a new value. However, the temperature stability is not as high as that of a
static insert. It is also possible to block the small capillary that feeds the cryogen to the
heat exchanger with frozen water or air during the sample changing operation if care is
not taken.
4.4.2
Static systems
Static systems are also fitted with heat exchangers, and the temperature of the heat
exchanger is controlled in a similar way. However, the exhaust gas does not flow over the
sample, but it passes out of the cryostat to the pump through a separate pumping line.
The heat exchanger usually forms an annulus around the sample space, and thermal
contact is made to the sample through exchange gas. The exchange gas pressure can be
adjusted to suit the conditions. The sample temperature follows the temperature of the
heat exchanger, but rapid temperature fluctuations tend to be filtered out, and the
temperature stability of the sample can be improved considerably. In some cases, a heater
is fitted to the sample block for fine control of the temperature or to warm the sample
quickly.

Page 29
29
Static inserts are as easy to operate as the dynamic type, and have the advantage that it is
not possible to block the heat exchanger during the sample changing process. Indeed,
quite large amounts of air may be frozen into the sample space without affecting the
operating procedure. However, the increased sample temperature stability has to be
traded off against the increased time taken to change the sample temperature to a new
value. In particular, it is not possible to cool the sample as quickly, and static systems are
generally used for small sample spaces.
Figure 9 Flow system for a continuous flow cryostat
4.5
Storage/transport dewars
Storage (or transport) dewars are generally only suitable for supplying cryogens to the
cryostat, (whether it is of the bath or continuous flow type). They are designed to be
robust and to have a low evaporation rate. They usually have very narrow necks and a
large amount of superinsulation. A few liquid helium storage dewars are fitted with
liquid nitrogen jackets (especially older dewars).
However, some small variable temperature inserts are available to fit into storage dewars,
providing that the diameter of the neck is sufficiently large, (50 mm). In particular, Oxford
Instruments can supply a variable temperature inserts (the Compact VTI), a
3
He refrigerator
(Heliox 2VL) or a
3
He/
4
He dilution refrigerator insert (Kelvinox15) to fit into a storage
dewar. These inserts give temperature ranges from 0.03 to 300 K.

Page 30
30
4.6
Closed cycle coolers
Modern closed cycle coolers offer a highly reliable method of achieving low temperatures.
They may either be used alone, to cool a sample and a radiation shield, or with a bath
cryostat to cool one or two radiation shields and thus reduce the evaporation rate of the
cryostat. This can considerably extend the hold time of a low loss cryostat, but it is not
usually appropriate if the equipment inside the cryostat has a high consumption rate
which has a dominant effect on the hold time. It is now possible to build cryogen free
systems containing superconducting magnets.
However, this type of cooler has a high initial cost and the pay back time (in terms of
reduced cryogen costs) may be very long. They also need to be serviced regularly (typically
every 5,000 hours). There is also the possibility of introducing unwanted vibration into the
experiment if it is not mounted very carefully.
4.7
'Stinger' systems
Some closed cycle cooler systems are used to re-condense helium gas into a bath cryostat
continuously. They take the form of a cold finger that fits into the helium reservoir. They
need quite high cooling powers both at the 4.2 K stage and at higher temperatures
because they have to provide enough cooling to replace the enthalpy of the boil off gas,
which usually helps to cool the neck of the reservoir. The helium reservoir is normally
pressurised slightly so that the gas recondenses effectively, and so the liquid helium is held
at a temperature close to 4.5 K.
4.8
Peltier effect coolers
Peltier effect coolers work by the thermoelectric effect; they are a thermodynamically
reversible low impedance devices, operating at a high current from a d.c. power supply. A
single stage cooler can typically achieve a temperature of -40°C, and lower temperatures
can be achieved using several stages. A six stage device may achieve -100°C and give a
cooling power of around 1 mW at -80°C. They do not introduce vibration into the
cryostat. Although they have a small temperature range and limited cooling power, they
offer a cheap solution for some requirements, (for example, Peltier effect cooled baffles,
see section 2.1.3).
4.9
Making indium seals
Oxford Instruments uses two main types of indium seal, as illustrated in Figure 21 on page
80. They both use 1mm diameter wire, retained
Either in a groove by a flat surface
Or in a corner between two flanges
In both cases, the indium wire is overlapped by bending one end of the wire sharply
outwards and laying the other end across the corner of the bend. The wire is so soft that
the joint will be compressed into a cold weld.

Page 31
31
4.9.1
Preparations
Before you make the seal ensure that the groove and the mating surfaces are clean.
Thoroughly remove any old indium wire from the seal faces. If necessary a solvent can be
used for cleaning. Some people like to grease the metal surfaces with silicone vacuum
grease to make it easier to remove the wire later, but this is not necessary.
4.9.2
Making the seal
Lay a new piece of indium wire in the groove or round the male spigot on one of the
flanges and overlap it as shown on the diagram. There are usually alignment marks on
the flanges to indicate the correct orientation. Carefully bring the two flanges together
and hold them loosely in place with two bolts while you put the other bolts into the
flanges and tighten them by finger only. Slowly and evenly tighten all of the bolts with a
small spanner (wrench) or Allen key. Do not tighten them too much. There is no need to
use an extension on the tool to give extra leverage. On large seals (typically > 50mm
diameter) it is then best to leave them for about an hour. The indium flows slightly
during this period so it is often possible to tighten the bolts slightly more.
4.9.3
Separating indium seal flanges
It is often difficult to separate indium seal flanges because the indium metal seems to glue
them together. Most large indium seals made by Oxford Instruments have two or more
threaded holes in one of the flanges for 'jacking screws'.
Remove the bolts that hold the indium seal together (leaving two of the bolts loosely in
place so that the flanges do not fall apart when they separate). Use another two of these
bolts to jack the flanges apart by screwing them evenly into the jacking screw holes from
the same side of the flange. This will push the flanges apart.
If there are no jacking screw holes (as often happens on small diameter indium seals), the
flanges can be separated by inserting a sharp blade between the flanges. Make sure that
the blade does not slip and cut you as the flanges separate.

Page 32
32
5
Ultra low temperatures
Refrigerators working at temperatures below 1 K are used for a surprisingly diverse range
of applications in research establishments. A range of specialised techniques is used to
achieve these temperatures. Most of the systems described in the previous chapters use
liquid helium and liquid nitrogen to reach and maintain low temperatures, but it is
difficult to achieve temperatures significantly below 1 K using these cryogens alone.
However, most ultra-low temperature systems are immersed in liquid helium (
4
He) at 4.2 K,
so that the heat load from the surroundings is minimised.
It is possible to reach temperatures slightly below 1 K by pumping liquid
4
He to a low
pressure but very large pumps are required and it is not usually economically viable.
4
He
may also be used to give very low cooling powers at temperatures down to 0.7 K in 'vortex
refrigerators' which rely on the special properties of superfluid
4
He.
However, the valuable lighter isotope of helium,
3
He, is usually used in refrigerators
working below 1 K. Evaporating
3
He is used in some systems, and temperatures slightly
below 0.3 K can be achieved by reducing its vapour pressure. Temperatures below 0.3 K
are usually reached by continuously diluting a flow of
3
He in liquid
4
He using a
3
He/
4
He
dilution refrigerator.
5.1
3
He Refrigerators
3
He refrigerators are usually designed for routine operation in the temperature range
from 0.3 to 1.2 K, and they use evaporating
3
He as the refrigerant. Their operating range
can often be extended to 100 K or higher. Some of these systems can run continuously,
returning the liquid
3
He to the system to replace the evaporated liquid. Others work in
'single shot' mode, by pumping on a small charge of liquid
3
He condensed into the system.
In an efficient cryostat a 20 cm
3
charge of liquid
3
He may last for longer than 50 hours.
Small laboratory refrigerators may give a cooling power of a few milli-watts at 0.5 K, but
very large and high powered machines can give cooling powers of several watts at this
temperature.
5.1.1
Sorption pumped
3
He systems
Sorption pumped
3
He systems are usually single shot refrigerators, capable of high
performance operation for a limited time. Several types of system are available to suit the
majority of laboratory requirements. Most of them can be used with high field
superconducting magnets if required. The top loading systems allow the sample to be
mounted on a probe which is loaded directly into liquid
3
He. They may also be designed
to operate in rapidly sweeping magnetic fields, and a wide range of special services may
be fitted to make connections to the sample. The maximum temperature limit is typically
100 K.

Page 33
33
The Heliox 2VL insert is a low cost miniature
3
He system designed to allow inexperienced
users to cool samples to 0.3 K. It is designed for operation in a liquid helium storage
dewar, or with a superconducting magnet system. The sample is mounted in vacuum, and
wiring can be connected easily. The whole insert is removed from the cryostat to change
the sample, but since it is small, the time scale for sample changing is similar to that on the
top loading systems. The Heliox system can be run up to about 200 K if it is used with a
superconducting magnet, but higher temperatures (up to 300 K) can be reached if the
insert is pulled up into the neck of the cryostat.
Figure 10 Principle of operation of a typical sorption pumped
3
He system
(top loading type)
Figure 10 shows the working parts of a typical system. Although a top loading insert is
shown, the principle of operation is similar for all Oxford Instruments' sorption pumped
inserts. The insert has an inner vacuum chamber, (IVC), to provide thermal isolation from
the main liquid helium bath.
The sorption pump, (or sorb), will absorb gas when cooled below 40 K, and the amount of
gas that can be absorbed depends on its temperature. It is cooled by drawing some liquid
helium from the main bath through a heat exchanger. The flow of
4
He through the heat
exchanger is promoted by a small diaphragm pump and the rate of flow is controlled by a
valve in the pumping line. A heater is fitted to the sorb so that its temperature can be
controlled.
The 1 K pot is used to condense the
3
He gas and then to reduce the amount of heat
conducted to the sample space. It is fed from the main liquid helium bath through a
needle valve, and it can be filled continuously.

Page 34
34
During condensation, the sorb is warmed above 40 K. When it is at this temperature it will
not absorb any
3
He (see Figure 10). The
3
He condenses on the 1 K pot assembly and runs
down to cool the sample and
3
He pot to the temperature of the 1 K pot. When most of
the gas has condensed into the insert, the 1 K pot needle valve is closed completely so that
the pot cools to the lowest possible temperature for optimum condensation. At this stage
the
3
He pot is full of liquid
3
He at approximately 1.2 K. The sorb is now cooled, and it
begins to reduce the vapour pressure above the liquid
3
He, (see Figure 10), so the sample
temperature drops. As the limiting pressure is approached, the temperature of the liquid
3
He can be reduced to below 0.3 K.
The temperature of the sample can be controlled by adjusting the temperature of the
sorb. If the sorb temperature is set between 10 and 40 K it is possible to control the
pressure of the
3
He vapour, and thus the temperature of the liquid
3
He. However, if the
best stability is needed, a temperature controller can be set up to measure the sample
temperature and control the power supplied to the sorb heater. No heat is supplied
directly to the liquid
3
He; this would evaporate it too quickly. The temperature of the sorb
is continuously adjusted by the temperature controller, and the temperature of the sample
can typically be maintained within 1 mK of the set temperature for the full hold time of
the system.
These systems have limitations both in their cooling power and base temperature, and if
high cooling powers (> 5 mW) are required, or operation must be continuous, it may be
more appropriate to choose a continuously circulating
3
He refrigerator. If however, the
base temperature is not low enough, a dilution refrigerator should be chosen. In general
it is found that a dilution refrigerator has a better performance below 0.4 to 0.5 K, and a
continuous
3
He system is better above this temperature. In either case, these refrigerators
typically have large room temperature pumping systems, and they are therefore rather
more expensive.
5.1.2
Continuously circulating
3
He refrigerators
Continuously circulating
3
He refrigerators are capable of giving high cooling powers and
of operating continuously for a long period. They use an external room temperature
pumping system (including a rotary pump and a booster pump).
The
3
He gas is injected into the cryostat and it is cooled to approximately 4.2 K by the
liquid helium bath before it enters the IVC. It is then cooled to 1.2 K and condensed by
the 1 K pot.
The liquid
3
He then passes through a special heat exchanger where it is cooled by the
outgoing
3
He gas. Below this heat exchanger an impedance is used to keep the pressure
in the condenser high enough even if the pressure in the
3
He pot is very low. On some
systems a needle valve is used here as a variable impedance to set the
3
He flow rate. Since
the liquid has already been cooled to a temperature close to that of the
3
He pot in the
heat exchanger, only a small fraction of it evaporates as it expands through the needle
valve. This ensures that the maximum amount of latent heat is available from a given
flow rate of
3
He. The liquid and gas then enters the
3
He pot, which has a large surface
area to give good thermal contact to the sample.

Page 35
35
The flow rate determines both the base temperature and the cooling power available
from the system. In general, a low flow rate will be required for a good base
temperature, and a high flow rate will allow a high cooling power to be achieved.
5.2
3
He/
4
He Dilution refrigerators
The principle of operation of the dilution refrigerator was originally proposed by
H. London in 1951, but the first working systems were not built until more than ten years
later. Since that time, the performance of these systems has steadily improved, and the
physical processes involved have become much better understood.
When a mixture of the two stable isotopes of helium is cooled below a critical
temperature it separates into two phases. The lighter 'concentrated phase' is rich in
3
He
and the heavier 'dilute phase' is rich in
4
He. The concentration of
3
He in each phase
depends upon the temperature. Since the enthalpy of the
3
He in the two phases is
different, it is possible to obtain cooling by 'evaporating' the
3
He from the concentrated
phase into the dilute phase.
Figure 11 Phase diagram of
3
He/
4
He mixtures
The properties of the liquids in the dilution refrigerator are described by quantum
mechanics and the details will not be described here. However, it is helpful to regard the
concentrated phase of the mixture as liquid
3
He, and the dilute phase as
3
He gas. The
4
He
which makes up the majority of the dilute phase is inert, and the
3
He 'gas' moves through
the liquid
4
He without interaction. This 'gas' is formed in the mixing chamber at the phase
boundary. This process continues to work even at the lowest temperatures because the
equilibrium concentration of
3
He in the dilute phase is still finite, even as the temperature
approaches absolute zero. However, the base temperature is limited by other factors, and
in particular by the residual heat leak and heat exchanger performance.

Page 36
36
Figure 12 Schematic diagram of a dilution refrigerator

Page 37
37
When the refrigerator is started the 1 K pot is used to condense the
3
He/
4
He mixture into
the dilution unit. It is not intended to cool the mixture enough to set up the phase
boundary but only to cool it to 1.2 K. In order to get phase separation, the temperature
must be reduced to below 0.86 K (the tri-critical point). The still is the first part of the
fridge to cool below 1.2 K. It cools the incoming
3
He before it enters the heat exchangers
and the mixing chamber, and phase separation typically occurs after a few minutes.
Gradually, the rest of the dilution unit is cooled to the point where phase separation
occurs.
It is important for the operation of the refrigerator that the
3
He concentration and the
volume of mixture is chosen correctly, so that the phase boundary is inside the mixing
chamber, and the liquid surface is in the still. The concentration of
3
He in the mixture is
typically between 10 and 20%.
In a continuously operating system, the
3
He must be extracted from the dilute phase (to
prevent it from saturating) and returned into the concentrated phase keeping the system
in a dynamic equilibrium. Figure 12 shows a schematic diagram of a typical continuously
operating dilution refrigerator. The
3
He is pumped away from the liquid surface in the
still, which is typically maintained at a temperature of 0.6 to 0.7 K. At this temperature
the vapour pressure of the
3
He is about 1000 times higher than that of
4
He, so
3
He
evaporates preferentially. A small amount of heat is supplied to the still to promote the
required flow.
The concentration of the
3
He in the dilute phase in the still therefore becomes lower than
it is in the mixing chamber, and the osmotic pressure difference drives a flow of
3
He to the
still. The
3
He leaving the mixing chamber is used to cool the returning flow of
concentrated
3
He in a series of heat exchangers. In the region where the temperature is
above about 50 mK, a conventional counterflow heat exchanger can be used effectively,
but at lower temperatures than this, the thermal boundary resistance (Kapitza resistance)
between the liquid and the solid walls increases with T
-3
, and so the contact area has to be
increased as far as possible. This is often done by using sintered silver heat exchangers,
which are very efficient even at the lowest temperatures.
The room temperature vacuum pumping system is used to remove the
3
He from the still,
and compress it to a pressure of a few hundred millibar. The gas is then passed through
filters and cold traps to remove impurities and returned to the cryostat, where it is pre-
cooled in the main helium bath and condensed on the 1 K pot. The primary impedance is
used to maintain a high enough pressure in the 1 K pot region for the gas to condense.
The experimental apparatus is mounted on or inside the mixing chamber, ensuring that it
is in good thermal contact with the dilute phase. All connections to the room
temperature equipment must be thermally anchored at various points on the refrigerator
to reduce the heat load on the mixing chamber and give the lowest possible base
temperature. If the experiment is to be carried out at higher temperatures, the mixing
chamber can be warmed by applying heat to it directly, and a temperature controller can
be used to give good stability.

Page 38
38
5.3
Sorption pumped dilution refrigerators
It is possible to build continuous dilution refrigerators which do not have external pumps
for the
3
He/
4
He mixture. Instead, two sorption pumps are used to pump the still to a low
pressure. A cold valve is fitted between each sorb and the still. While one of the sorbs is
pumping, the other is regenerating. The temperatures of the sorbs are adjusted by
electrical heaters to control the pumping cycle.
A special 'collector' is fitted below the 1 K pot to hold the liquid condensed by the pot.
The pressure in this collector is controlled by maintaining a constant temperature, so that
the flow of
3
He to the dilution unit is kept constant even though the flow from the pumps
to the condenser is not constant.
The advantages of these systems are that the vibration levels can be significantly reduced,
and the refrigerator system is compact. Since the
3
He/
4
He mixture remains in the cryostat it
is less likely that air can leak into it and block the system. They are controlled by a
computer, so they can be automated easily.
5.4
Nuclear demagnetisation systems
Temperatures below approximately 4 mK cannot be achieved easily or cheaply. Dilution
refrigerators capable of reaching temperatures below 5 mK are available but they are
large and expensive. Although temperatures as low as 2 mK have been achieved in this
type of system, most experimentalists use other techniques.
Most experiments carried out below 4 mK rely on adiabatic demagnetisation of a nuclear
paramagnet. This is a single shot process, but very long hold times can be achieved.
However, the total amount of heat that can be absorbed from the sample by the
demagnetisation stage is limited. Demagnetisation stages are typically pre-cooled to
approximately 10 mK in a magnetic field of 8 to 10 T by a powerful dilution refrigerator.
They are then isolated from the mixing chamber by a superconducting heat switch, and
the magnetic field is slowly reduced. Temperatures slightly below 1 mK can be achieved
using PrNi
5
(an enhanced nuclear paramagnet), but copper can be demagnetised to
around 10 µK.

Page 39
39
6
Superconducting magnet technology
6.1
Introduction
The world's first commercial superconducting magnet was produced by Oxford
Instruments., and now, more than 25 years later the company still leads the world, with
fields higher than 20 T available. This technology allows customers to produce extremely
high magnetic fields in laboratory scale cryostats without the kW to MW power supplies
needed for non-superconducting magnets. In most cases the cost of refrigeration for a
superconducting system is much less than the cost of the power required to run an
equivalent non-superconducting system.
Many types of magnet are available, but solenoids and split pairs (sometimes referred to
as split solenoids) are the types most commonly encountered in the laboratory. These two
types of magnet are shown schematically in Figure 13. Solenoids are generally simpler,
and it is cheaper to produce a magnet with a given field using a solenoid than it is using a
split pair. It is also generally possible to achieve better homogeneity of the magnetic field
using a solenoid. The very high forces between the coils make it difficult to produce fields
higher than 15 T using a split pair magnet. However split pairs give access to the sample
perpendicular to the magnetic field. They are commonly used for optical experiments
which require this access.
Fields up to 9 T are usually produced using NbTi superconductor at 4.2 K (or 11 T at 2.2 K);
higher fields (up to 20 T) require the use of the expensive and brittle intermetallic
compound Nb
3
Sn. However the Nb
3
Sn is only used for the inner sections of such a magnet
(where the field is highest) and the outer sections use the cheaper NbTi for economic
reasons. Many kilometres of wire are used in the winding of even a modest magnet.
Fields up to about 40 T can be achieved with hybrid magnets. In this type of system a
large bore superconducting magnet provides the background field (up to 16 T) for a high
power, water cooled inner winding.
It is also likely that 'high T
C
' superconducting inner coils will soon be commercially
available to enhance the field produced by a Nb
3
Sn magnet. Although these materials
cannot yet tolerate very high current densities they have exceptionally high critical fields
when cooled to 4.2 K.
Additional coils may be fitted to the basic windings to modify the shape of the field.
'Compensation coils' are often used to improve the homogeneity at the centre of field by
reducing the rate at which the field drops at the ends of the coils (due to finite winding
length effects). 'Shim coils' (or shims) are used to remove residual field gradients; they
may be wired in series with the main coils to give a basic level of correction or
independently to give finer adjustment. Shims may be either cold superconducting coils
or room temperature 'normal' coils.
'Cancellation coils' are often fitted to one end (or sometimes both ends) of a magnet to
give a low field region quite close to the centre of field; for example, < 10 mT (or
100 gauss) may be achieved over a region only 30 cm away from the centre of field of a
15 T magnet.

Page 40
40
Figure 13 Schematic diagram of a simple solenoid and split pair
6.2
Construction of the magnet
Superconducting magnets are typically constructed from a number of coaxial coils. They
are wound from different grades of superconductor so that the cost is reduced as far as
possible. The coils are impregnated to give a high mechanical stability and thus to prevent
relative movement of the components as the field is changed. The Oxford Instruments
'Magnabond' system has been developed to achieve this.
The coils of split pair magnets have to be supported especially carefully to resist the large
forces between and within the coils. These forces are typically tens of tonnes.
Compensation coils, shim coils and cancellation coils also have to be fixed very firmly to
the main coils.
Electrical connections between the coils have to be made using superconducting joints so
that the residual resistance of the magnet is reduced to a minimal level, (typically lower
than 10
-8
ohms).
The wire used in the construction of most laboratory magnets is multi-filamentary,
because this improves the stability by preventing 'flux jumping' which dissipates energy in
the superconductor. However, some low decay rate magnets are wound using single core
wire because it is possible to make lower resistance joints in conductors of this type,
(typically 10
-14
ohms or lower).

Page 41
41
6.3
Basic physics of the magnet
Although the basic physics taught to a 16 year old is sufficient to explain many of the
phenomena observed in a magnet, the production of a reliable magnet is extremely
challenging. The magnet is effectively a pure inductor with zero resistance, and the circuit
theory taught in schools explains that the magnet stores energy, that there is a time
constant associated with a circuit containing an inductor and a resistor in parallel, and that
it is difficult to change the current flowing in the inductor because of the induced back
e.m.f. (or voltage).
Stored energy
LI
=
1
2
2
Induced back e m f
L
dI
dt
'
. . '.= −
Time constant
L
R
=
where L is the inductance of the magnet, I is the current in the magnet, and R is the
resistance in parallel with the windings. As an example, a magnet with an inductance of
100 H is not unknown in a laboratory cryostat, and if it was operating at a current of
100 A, the stored energy would be ½ MJ!
The induced (or 'inductive') voltage observed when the magnet is energised or de-
energised is explained further in the section on the typical operation procedure (see
section 6.9 on page 46). In many cases this induced voltage limits the rate at which the
magnetic field can be changed (or 'swept'), because of the limitations of the power
supply. However, there are several effects within the windings of the magnet that cause
heating (for example, eddy currents, hysteresis and diamagnetism), and ultimately these
limit the sweep rate.
Large stresses are induced in the windings of the magnet because of the Lorentz forces
between the field and current. These forces lead to large hoop stresses (trying to explode
the magnet) and axial compression in the windings.
For simple operation the magnet is cooled to 4.2 K using liquid helium at its normal
boiling point. However, the properties of the superconducting materials in the windings
of the magnet are improved when their temperature is reduced further. In many cases it
is possible to obtain an enhanced performance by cooling the magnet to 2.2 K and
energising it to a higher current. Enhancements of the order of 20 to 25% can typically be
achieved. However, it is important to check that the magnet is designed to withstand the
increased stresses before attempting to run it in this way, otherwise it may be badly
damaged. The temperature is commonly reduced by a 'lambda point refrigerator', as
described in section 4.2 on page 24.

Page 42
42
Ferromagnetic materials close to the magnet can have positive or negative effects. On the
positive side, they may be used for fine 'shimming' of the magnetic field in certain
specialised magnets, or more commonly to reduce stray magnetic field to an acceptable
level. On the negative side, they may make unwanted changes to the field shape in a
region of high homogeneity and put extremely high forces on the magnet windings or the
cryostat. The additional stress on the windings may even prevent the magnet from
functioning correctly. For these reasons, any large magnetic items have to be positioned
carefully, so that they neither affect the field shape nor cause damage to the magnet. If
shielding is required, the effects on the magnet must be analysed carefully by computer
simulation, and this is such a specialised field that it should only be undertaken by an
expert.
6.4
Homogeneity of the field
The homogeneity of the magnetic field is often specified over a 10 mm diameter spherical
volume (or d.s.v.). In a solenoid type magnet, a homogeneity of 1 in 10
3
can easily be
achieved, and this is sufficiently high for the majority of experiments. This can be
improved to 1 in 10
5
by using series shims (sufficient for low resolution NMR). However,
high resolution NMR and similar experiments are usually carried out in magnets with
homogeneity 1 in 10
7
(or better) over a 10 mm d.s.v., which can only be achieved using
independent shims. It is much more difficult to obtain high homogeneity in a split pair
magnet, and 1 in 10
2